hy Everybody:
we hav a sytem PQ2000,when v on the sytem at initital time it does,n,t peroform warmup showing errror Bad inverter" shoot thru" bad Ac in " etc.
when we increase tube heat using
manexp:( low kv, low ma, no rotation,) it perform scanning, after pre-heat completing,then system perform Good in all days.
when v off the system then on, it showing error again,
is it,s the reason of XSC?(XSC does,n,t save preheat calculation)
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
munsi khan
help on PQ 2000
July 05, 2009 09:52
hy Everybody:
we hav a sytem PQ2000,when v on the sytem at initital time it does,n,t peroform warmup showing errror Bad inverter" shoot thru" bad Ac in " etc.
when we increase tube heat using
manexp:( low kv, low ma, no rotation,) it perform scanning, after pre-heat completing,then system perform Good in all days.
when v off the system then on, it showing error again,
is it,s the reason of XSC?(XSC does,n,t save preheat calculation)
thanks in Advance
to rate and post a comment
Sergio Leonardo Barral
Re: help on PQ 2000
July 09, 2009 10:13
Munsi:
It's difficult to me to understand what happen to your PQ. Can you write the complete error message?
BEst regards
Sergio
to rate and post a comment
Alexander Tonev
Re: help on PQ 2000
July 10, 2009 01:35
Replace DS1244Y