If this Unit has anatomical programing,when they select this shot the bucky may be configed wrong,perhaps selecting table instead of wall bucky.Is this the only shot your getting grid lines on, is this the only shot they are visable. maybe grid is not moving at all.
From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729
Marvin Morales
Help Picker HF 500
August 23, 2010 10:04
Hello
I have problem with grid. the technician gets the grid lines on the images whe the exposure is for head or neck areas.
to rate and post a comment
Doug Gray
Re: Help Picker HF 500
August 31, 2010 11:06
Does the unit have a bucky, or a grid cabnet?
to rate and post a comment
Doug Lally
Re: Help Picker HF 500
November 30, 2011 10:49
I think this hf 500 is really a MP 500 made by Universal and picker slapped their name on it. Call Del medical ask for Tony Anitch in tech support.
to rate and post a comment
Japhet Kamala
re: Re: Help Picker HF 500
August 06, 2013 11:16
Dear All
Kindly am seeking picker HF 500 generator x ray manual
some one assist me
to rate and post a comment
Doug Lally
Re: Help Picker HF 500
December 18, 2011 11:31
If this Unit has anatomical programing,when they select this shot the bucky may be configed wrong,perhaps selecting table instead of wall bucky.Is this the only shot your getting grid lines on, is this the only shot they are visable. maybe grid is not moving at all.