Eugene Ismagilov

Help.Error code 524. Picker CT PQ5000

March 18, 2020 11:22

Hello,
I need help with 524 error code for Picker
PQ5000 CT scanner. Warmup and preheats
run ok, but any scan doesnot work.
I have already replaced xsc, sif,brushes and
so on with no success. Please help with
advice.

Thanks.
Eugene Ismagilov

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: Help.Error code 524. Picker CT PQ5000

March 20, 2020 10:41

524 - XSC state error. The ASCII message contains the device in error along with the expected state. The actual
device state is in the SIF error buffer.
Check XSC STAT: 2nd LSB nibble, Bit 4 for boot status. If set, check slip rings 3 & 4 (New style's Fuses)

Log inor Register

to rate and post a comment

(55)
(1)
(1)

gustavo fiochi

re: Help.Error code 524. Picker CT PQ5000

March 24, 2020 11:06

524 is an error from xcs but has many causes. see erp and ask for long form to see more details. email me if you don t know how

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

re: Help.Error code 524. Picker CT PQ5000

March 27, 2020 10:25

Hi
First. If you replaced the signal brush board, VERIFY continuity between pins 3 and 4 are correct. Some of these boards have a different color connector and the polarization pin in the female connector on board is located on the opposite side causing you to place the cable in the wrong position. it never will communicate unless you correct the polarization pin position.
If the original error was originated due to a bad signal brush board, the GPU board has the DUART ports blown and you need to replace them or the whole board.
Your problem is one of these 2 option
Best regards

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)