Hello,
Physicists found my Litho unit has a Hi Contrast resolution issue. Here's the following found by physicist : Minimum Resolution = 2lp/mm x (15cm/(FOV in cm))

1) 18cm FOV requires 1.7lp/mm - UMP observed 1.4lp/mm ***( I can verify 1.5lp/mm with a Gammex LinePair Phantom)

2) 12cm FOV requires 2.5lp/mm - UMP observed 2.0lp/mm ***( I can verify 2.1lp/mm with a Gammex Linepair Phantom)

What can I adjust or will have to replace a part? I have no service manuals for this unit.
Thanks for any help.

Jere

Log inor Register

to rate and post a comment

Artemio Santacruz

re: Hi Contrast Resolution issue

October 30, 2014 11:51

try put the resolution phanton in fromm off II and try to ajuste g1 ,g2 ,looking the quality in the monitor

Log inor Register

to rate and post a comment

(12)
(6)

Kaustubh Dere

re: Hi Contrast Resolution issue

October 30, 2014 11:52

Can u tell me which model of Lithotripter are you referring

Log inor Register

to rate and post a comment

(17)
(1)

Jeremiah Kiblin

re: re: Hi Contrast Resolution issue

October 30, 2014 04:36

Siemens Lithotriper - MODULARIS, SIREMOBILE ICO-C

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Hi Contrast Resolution issue

October 30, 2014 11:52

which unit is it. normally this could be a sign of a bad II but it could also be a calibration issue. I normally perform a complete generator calibration then calibrate the doses and recalibrate the image chain.
If you let me know what system it is I might be able to help you with the manuals

Log inor Register

to rate and post a comment

(17)
(1)

Jeremiah Kiblin

re: re: Hi Contrast Resolution issue

October 30, 2014 04:36

Siemens Lithotriper - MODULARIS, SIREMOBILE ICO-C

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Hi Contrast Resolution issue

October 31, 2014 08:57

I can send you the manuals if you give me an email address.

Log inor Register

to rate and post a comment

(17)
(1)

Jeremiah Kiblin

re: re: re: re: Hi Contrast Resolution issue

October 31, 2014 09:00

jeremiah.kiblin@nfmmc.org and Thank you

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: Hi Contrast Resolution issue

October 31, 2014 05:13

I will upload and send it to you

Log inor Register

to rate and post a comment

Artemio Santacruz

re: re: Hi Contrast Resolution issue

October 31, 2014 08:58

somme time inside off II the higth voltage power suplies after the timer the resistor change de value normally biger value or open you can mesure

Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: Hi Contrast Resolution issue

October 30, 2014 11:53

Basically Hi contrast resolution is focus. First check to see if your camera has a lens. If so then you can try adjusting the optical focus while observing a resolution pattern during fluoro. There may be an electrical focus adjust on the camera also. These adjustments will affect over all focus on all FOV so if you improve the 12 cm FOV it should also improve the MAG. There are focus adjustments on the II power supply for each mode also. But I would try the camera end first.

Log inor Register

to rate and post a comment

(17)
(1)

Jeremiah Kiblin

re: Hi Contrast Resolution issue

October 30, 2014 04:36

Anyone have a manual or you can send me an image of where the adjustments are located from manual to jeremiah.kiblin@nfmmc.org ? I have no drawings or schematics on this Lithotriper.

Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: re: Hi Contrast Resolution issue

October 31, 2014 05:14

What is the manufacturer of lithotripter. What kind of C-arm?