Hello every body, I have an issue with Highspeed Dual CT scan. The issue is that no power on OC, when you power the equipment, only tube fan start running.
We have found 2 burns resistors on RMT BD ASSY in the PDU, then we have replace the board but the problem persit.
24V input voltage is lights, K9 still lights, no light on other relays
Any help will be appreciated
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Youbissi Pascal
Highspeed DUAL Problem
April 21, 2014 11:44
Hello every body, I have an issue with Highspeed Dual CT scan. The issue is that no power on OC, when you power the equipment, only tube fan start running.
We have found 2 burns resistors on RMT BD ASSY in the PDU, then we have replace the board but the problem persit.
24V input voltage is lights, K9 still lights, no light on other relays
Any help will be appreciated
to rate and post a comment
Carlos Duran
re: Highspeed DUAL Problem
April 22, 2014 10:04
Check the breakers below RMT BD, one of them is for console power