Host: TAC Proc: tgp_out Error: 186518
File: tgp_errors.c Line:312
DAS Error.
Code: 30-1010-01
Error detected by: CIF
Zero DAS data was detected.
I've checked the DAS PS and re-adjusted the +12Vdc (from 12.36V to 12.46V) and -12Vdc (from -12.37V to -12.43V), I checked the CIF board Dip Switches (SW2 and SW3) and the internal switches are OFF in both of the Dip Switches.
Is there any other procedure to check if the CIF and the DDP boards are working good?
Does anyone have had this same issue, and want to share the solution with me?
FranciscoG
HighSpeed NX/i Error: Zero DAS data was detected
January 26, 2018 09:19
Hi guys, I'm having this issue with this system:
Host: TAC Proc: tgp_out Error: 186518
File: tgp_errors.c Line:312
DAS Error.
Code: 30-1010-01
Error detected by: CIF
Zero DAS data was detected.
I've checked the DAS PS and re-adjusted the +12Vdc (from 12.36V to 12.46V) and -12Vdc (from -12.37V to -12.43V), I checked the CIF board Dip Switches (SW2 and SW3) and the internal switches are OFF in both of the Dip Switches.
Is there any other procedure to check if the CIF and the DDP boards are working good?
Does anyone have had this same issue, and want to share the solution with me?
Thanks in advance for your time and thoughts.
to rate and post a comment
CRAIG WAMBURA
re: HighSpeed NX/i Error: Zero DAS data was detected
January 27, 2018 12:56
Would check RF antenna, slip ring and fibre optic
to rate and post a comment
john dismaya
re:HighSpeed NX/i Error: Zero DAS data was detected
January 28, 2018 11:25
Run DAS data transfer test. Try stationary ang rotating.