Problems with right side controls does not response.
LCD works fine and also right side controls are ok.
Any idea about cause of this problem? (I already check cable connections)
It's probably the right-side PCB assembly. You'll probably have to replace it. And if it's not the board, then it's the wiring harness between the board and the Junction Box. But have you gotten into the Service Mode of the TotalCare to verify the problem?
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Manuel Esquer
Hill-Rom P1900 controls problem
October 11, 2021 10:33
Problems with right side controls does not response.
LCD works fine and also right side controls are ok.
Any idea about cause of this problem? (I already check cable connections)
to rate and post a comment
Ray Brown
re: Hill-Rom P1900 controls problem
October 12, 2021 10:05
It's probably the right-side PCB assembly. You'll probably have to replace it. And if it's not the board, then it's the wiring harness between the board and the Junction Box. But have you gotten into the Service Mode of the TotalCare to verify the problem?
to rate and post a comment