Hi guys,

I've got the following errors on this HiSpeed:

1. Code: 20-1014-03
2. Code: 20-0014-06
3. Code: 11-10F2-01

All cables and connectors are checked, ant the system won´t let me even warm up the tube. Any help is appreciated.

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: HiSpeed Dual CT

February 11, 2017 03:39

1. 11-10F2-01 SAFETY LOOP OPEN.
Error mA/kV status was not received from OGP
during X-ray ON. Or X-ray OFF Status was not
received correctly from OGP. Safety Loop was
opened by TGP as the result.

JEDI(kV control orCT-IF) or interface between
JEDI and TGP including OGP and Slip Ring

2. 20-0014-06 CAN BUS Warning status.
Infor. Some retries happened on CAN bus between
JEDI and OGP and CAN bus becomes
WARNING Status as the result.

JEDI(kV control or CT-IF), or interface between
JEDI and OGP

3. 20-1014-03 No answer of mAkV. Error OGP did not receive reply from JEDI in response to mA/kV setup request.

JEDI(kV control or CT-IF), or interface between
JEDI and OGP

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: HiSpeed Dual CT

April 06, 2017 05:31

Thank you Karen, I've tested the OGP and TGP into another working system, and found out that the OGP was bad, so I replaced it, only now getting this messages when trying to warm up the tube:

Host: Linux Proc:tgp_out Error:186395
File:tgp_errors.c Line:376
XG Proc Comm Error.
Code:20-1014-03
Error detected by: OGP
OGP did not receive reply from JEDI in response to mA/kV setup request


Host: Linux Proc:tgp_out Error:186364
File:tgp_errors.c Line:453
XG Proc Comm Error.
Code:11-1010-02
Error detected by: TGP(MP)
TGP(MP) did not receive neither ACK nor NACK from 'OGP'


Host: Linux Proc:tgp_out Error:186221
File:tmup_collimator_request.c Line:79
Collimator Request Timer Timeup.


Host: Linux Proc:ScanCtrl Error:186117
File:tgp_input.c Line:2019
Error detected at TGP_COLLIMATOR_REQUEST_RESP.(code=-1)


Host: Linux Proc:tgp_out Error:186363
File:tgp_errors.c Line:376
Scan Proc Comm Error.
Code:11-1010-01
Error detected by: TGP(MP)
TGP(MP) received the message 'Not Acknowledged' from OGP more than 3 times

Any help is appreciated.

1 person liked this
Log inor Register

to rate and post a comment

(33)
(1)

sathish boina

re: re: re: HiSpeed Dual CT

April 07, 2017 10:22

Check input voltage to ogp, increase slightly if required,24v and 5v

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)