Air cal data all RED,You have a bad CAM BD.Plot the air cal data from the daily calibration,too see which BD is bad,or just trouble shoot it like any other das problem,YOU do know how to do that
thanks everybody for your help.I had perform the water calibration and air calibration.but it's no useful.
dear verne chappell,I don't know how to troubleshoot this problem.can you tell me how to do?thanks.
are their any of the service manuals on site,if so start reading,toohave a basic knowledge of the system.IF NOT go to this link at ge
www.gehealthcare.com/usen/service/biomed_tech_selfservice/services_user_doc/products/support.CLICK
on CT,then select the option CT/e,CT Prospeed AI/FI,select all and click on search.Then you can select the manuals you want and down load then.These are the basic service manuals.ZIP is faster,pdf takes some time.Email me direct at scanworks@terra.com.br.This link to GE is a good link,you can find it on their web site,but it takes some serching and time,Its a free link and they are providing it,NO SECRET STUFF,but very usefull. verne
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
yuan bo
Hispeed FX/I problem.
August 10, 2007 07:47
The display of all air drift is red.I had perform the air calibration.but it's not useful.Who could help me?thanks
best regards.
to rate and post a comment
Timothy Bascon
Re: Hispeed FX/I problem.
August 10, 2007 10:26
If air cal still wouldn't work, try restoring the whole calibration from the MOD (if you have it).
to rate and post a comment
VERNE CHAPPELL
Re: Hispeed FX/I problem.
August 11, 2007 08:00
Air cal data all RED,You have a bad CAM BD.Plot the air cal data from the daily calibration,too see which BD is bad,or just trouble shoot it like any other das problem,YOU do know how to do that
to rate and post a comment
yuan bo
Re: Hispeed FX/I problem.
August 12, 2007 06:49
thanks everybody for your help.I had perform the water calibration and air calibration.but it's no useful.
dear verne chappell,I don't know how to troubleshoot this problem.can you tell me how to do?thanks.
best regards
to rate and post a comment
VERNE CHAPPELL
Re: Hispeed FX/I problem.
August 13, 2007 07:25
are their any of the service manuals on site,if so start reading,toohave a basic knowledge of the system.IF NOT go to this link at ge
www.gehealthcare.com/usen/service/biomed_tech_selfservice/services_user_doc/products/support.CLICK
on CT,then select the option CT/e,CT Prospeed AI/FI,select all and click on search.Then you can select the manuals you want and down load then.These are the basic service manuals.ZIP is faster,pdf takes some time.Email me direct at scanworks@terra.com.br.This link to GE is a good link,you can find it on their web site,but it takes some serching and time,Its a free link and they are providing it,NO SECRET STUFF,but very usefull. verne
to rate and post a comment
Sanjeev Kumar
Re: Hispeed FX/I problem.
August 10, 2007 01:52
Dear,
kindly check power supply of the monitor detector which is under pre collimator and try to do water phantom callibration too.
to rate and post a comment
gibarcorp
re: Hispeed FX/I problem.
February 03, 2020 10:15
Hi, do you have any solution in this matter? i am experiencing the same problem now. what is your recommendation and findings? thanks
to rate and post a comment