Your Email:
Manuel Esquer
HiSpeed LXi
November 01, 2010 03:58
I got this error on a CT GE HiSpeed LXi XG Error phase is 5 code is 60 detail error code: 60-0321H Tube Spits error: spit retry failed (spit TDcomputing error) I do not have troubleshooting procedure for this error on manua. Any idea?
to rate and post a comment
yuan bo
Re: HiSpeed LXi
November 02, 2010 09:52
i have the troubleshooting procedure, and it includes the potential cause and troubleshooting guide. if you need it, contact me. obnauy@163.com
JULIO REYES
November 03, 2010 11:05
Hi Manuel: You must have more than 1 error message but the right information for error 60-0312H is: Error code: 60–0312H Message/explanation: ILR current not OK. The current in the serial resonant circuit of the inverter did not rise at the beginning of the exposure. Potential cause: 1. Inverter LC resonant circuit (Inverter coil assy, capa inverter assy.) 2. Inverter 3. kV control Recommended action/ Troubleshooting guide: * Run HV power diagnostics I hope this can help you to fix your machine. Regards Julio REYES
Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com
Manuel Esquer
HiSpeed LXi
November 01, 2010 03:58
I got this error on a CT GE HiSpeed LXi
XG Error phase is 5 code is 60 detail error code: 60-0321H Tube Spits error: spit retry failed (spit TDcomputing error)
I do not have troubleshooting procedure for this error on manua.
Any idea?
to rate and post a comment
yuan bo
Re: HiSpeed LXi
November 02, 2010 09:52
i have the troubleshooting procedure, and it includes the potential cause and troubleshooting guide. if you need it, contact me. obnauy@163.com
to rate and post a comment
JULIO REYES
Re: HiSpeed LXi
November 03, 2010 11:05
Hi Manuel:
You must have more than 1 error message but the
right information for error 60-0312H is:
Error code: 60–0312H
Message/explanation: ILR current not OK.
The current in the serial
resonant circuit of the inverter
did not rise at the
beginning of the exposure.
Potential cause:
1. Inverter LC resonant circuit (Inverter coil
assy, capa inverter assy.)
2. Inverter
3. kV control
Recommended action/ Troubleshooting guide:
* Run HV power diagnostics
I hope this can help you to fix your machine.
Regards
Julio REYES
to rate and post a comment