I got this error on a CT GE HiSpeed LXi
XG Error phase is 5 code is 60 detail error code: 60-0321H Tube Spits error: spit retry failed (spit TDcomputing error)
I do not have troubleshooting procedure for this error on manua.
Any idea?

Log inor Register

to rate and post a comment

(373)
(66)
(15)

yuan bo

Re: HiSpeed LXi

November 02, 2010 09:52

i have the troubleshooting procedure, and it includes the potential cause and troubleshooting guide. if you need it, contact me. obnauy@163.com

Log inor Register

to rate and post a comment

(5)

JULIO REYES

Re: HiSpeed LXi

November 03, 2010 11:05

Hi Manuel:

You must have more than 1 error message but the
right information for error 60-0312H is:

Error code: 60–0312H

Message/explanation: ILR current not OK.
The current in the serial
resonant circuit of the inverter
did not rise at the
beginning of the exposure.

Potential cause:
1. Inverter LC resonant circuit (Inverter coil
assy, capa inverter assy.)
2. Inverter
3. kV control

Recommended action/ Troubleshooting guide:
* Run HV power diagnostics

I hope this can help you to fix your machine.

Regards

Julio REYES


Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)