Aleksander Proof
Hitachi Avius TCD problem
April 01, 2014 07:53
Hi friends!
I have problem with Hitachi Avius on TCD programm.
In the study of the head unit does not draw blood vessels and gives only red and blue ripples. Other protocols work without problems. I did reset the default protocol for TCD, but it did not help. I put the sensor in another slot, but it did not help. I put the sensor on another machine and it works without problems on another machine. How to conduct tests and find out what's the problem?
to rate and post a comment
Aleksander Proof
re: Hitachi Avius TCD problem
April 02, 2014 07:59
Hi guys!
Today I checked this probe itself on two other machines and it showed the same interference. Conclusion: The sensor is faulty. Who knows how to fix it, please contact us.
to rate and post a comment