I have some problem about Hitachi Supria 16. So that we saw two error.
1. An abnormality occurred in the X-ray Generator. Please retry. Error code: 0002D800 (X D8)
2. The number of arcing recoveries has been exceeded. Please retry. Error code: 00022000 (X 20)
Could you please help me to solve this problem.

1 person liked this
Log inor Register

to rate and post a comment

(67)
(19)

Tolga GULUMSER

re: Hitachi CT

August 20, 2020 11:03

What is the slice count of x-ray tube? If it is high, then first suspicious xray tube. And check the hv generator and cables.

Log inor Register

to rate and post a comment

(3)

turfa jhjah

re: re: Hitachi CT

October 31, 2023 10:37

I have same problem

Log inor Register

to rate and post a comment

(17)
(1)

Root

re: Hitachi CT

August 20, 2020 11:05

1 - X D8 means that : Moniteur Communication error in the inverter unit
2 - X 20 : Arcing recovery counts over

Log inor Register

to rate and post a comment

(66)
(5)

Mr.CT.service

re: re: Hitachi CT

February 16, 2022 10:55

Hi,
May you share manual for re-install/adjust X-Ray tube Varex of Hitachi Supra 16?

Best regards

Log inor Register

to rate and post a comment

(15)

Waleed Ahmed

re: re: Hitachi CT

October 31, 2022 09:36

Hello Sir,

Hitachi CT-32 Supria have error Code: 00029300 (X 93) if you can help me to gate service manual to solve this problem so that I can repair the device
If you can provide me the service manual for the device, I will be grateful for your cooperationIf, please help Sir,

Log inor Register

to rate and post a comment

(8)
(1)

Tchouto donatien

re: Hitachi CT

February 21, 2024 10:55

change the tube

Log inor Register