DJIELA Flavien
HITACHI EUB 525
January 13, 2009 04:15
Dear all,
I have got a HITACHI ultrasound that the test partten does not feature on the screen even with the probes connected.
Please can you help me to solve this problem
Any help will be appreciated.
Best Regard.
DJIELA
to rate and post a comment
Tajinder Singh Bhatia
Re: HITACHI EUB 525
January 14, 2009 05:52
1. Remove all probes and check for "probe not connected" message.
2. Connect only one probe and try at different ports.
Regards
T S Bhatia
to rate and post a comment
DJIELA Flavien
Re: HITACHI EUB 525
January 14, 2009 09:29
thanks Mr Bhatia,
I have tried those tests already with no success. I even changed the probe and still i couldn't have the test pattern shown on the screen.
please i am waiting for your reply.
Best Regards,
DJIELA
to rate and post a comment
Tajinder Singh Bhatia
Re: HITACHI EUB 525
January 17, 2009 03:17
Hi,
I was out so could not reply. Anyways, It seems problem could be with power supplies to memory board.
regards
T S Bhatia
www.mriservices.co.in
to rate and post a comment
DJIELA Flavien
Re: HITACHI EUB 525
January 19, 2009 03:19
Hi,
Thanks for your reply. At the power supply stage, i have checked all the voltages and there are normal.
I think there should be a faulty board which is causing this problem. So, my problem is now to know which of the boards is the one. Below is the list of the boards in the ultrasound:
- LDEL
- SDEL 1
- SDEL 2
- CONT
- DOP
- FFT
- CFM 1
- CFM 2
- BWDSC
- CMEM
- CDISP
- MCPU
- MM
Regards.
DJIELA