Dear all friends,
I have Hitachi EUB 8500
I pluged it into 220v by mistake instead of 110v (volt) and the the device didn't turn on any more just monitor power turned on . I think power supply has problem but where in power supply I don't know
I was wondering if anyone faced with this problem to help me ?
I don't have EUB 8500 manual. Could you please let me know if some one has it ?
Hamidreza F
The network has a manual in EUB-5500, they are not fundamentally different. The power supply is a separate unit, disposed downstream unit. Usually the outside jumper, determine the incoming voltage or transformer winding signed.
Dear Alex,
Thanks for your response
I opened the supply part which is the heavy Box at botom of device and I looked at the pcb board but I didn't find any component damage physically
I wanna check connector voltage to find with pcb board don't make voltage
Fir this reason I need manual 8500
and I don't know which jumper set to 220
I have same problem with EUB 8500. I opened the power box, but it is totally different from the EUB 5500. Also, I read the manual of 5500 and did not find any useful thing. Let me know what should I do, if you can help me.
I have found a table for CN5 at page 82 as follows:
Connector
No.
Input voltage Connection
100V /220V A1-B2, A2-B1 short
CN5 110V /230V A1-A3, A2-B1 short
120V /240V A1-B3, A2-B1 short
but, I really don't know what should I do with this. The connection of eub 8500 is different from 5500 and I have already connected a 110 volt hitachi 8500 to 220 volt and now it does not work.
I would be greatly tankful, if you clearly tell what exactly I should check.
We are the best choice for all your Endoscopy needs. We have Certified Pre-Owned refurbished and used Endoscopes, and all other related components and
accessories. We are committed to your absolute satisfaction. (513) 469-7267
Hitachi EUB 8500
July 29, 2015 07:57
Dear all friends,
I have Hitachi EUB 8500
I pluged it into 220v by mistake instead of 110v (volt) and the the device didn't turn on any more just monitor power turned on . I think power supply has problem but where in power supply I don't know
I was wondering if anyone faced with this problem to help me ?
I don't have EUB 8500 manual. Could you please let me know if some one has it ?
Hamidreza F
to rate and post a comment
Alex Mykhaylyk
re: Hitachi EUB 8500
July 29, 2015 10:29
The network has a manual in EUB-5500, they are not fundamentally different. The power supply is a separate unit, disposed downstream unit. Usually the outside jumper, determine the incoming voltage or transformer winding signed.
to rate and post a comment
Hamidreza fouladi
re: re: Hitachi EUB 8500
July 29, 2015 12:19
Dear Alex,
Thanks for your response
I opened the supply part which is the heavy Box at botom of device and I looked at the pcb board but I didn't find any component damage physically
I wanna check connector voltage to find with pcb board don't make voltage
Fir this reason I need manual 8500
and I don't know which jumper set to 220
This is my Gmail:
hamidreza.fooladi@gmail.com
to rate and post a comment
Alex Mykhaylyk
re: re: re: Hitachi EUB 8500
July 30, 2015 03:46
EUB-5500 Manual, page5-5 or 82,connector CN5.
to rate and post a comment
Ali Deylami
re: re: re: re: Hitachi EUB 8500
September 08, 2015 07:57
I have same problem with EUB 8500. I opened the power box, but it is totally different from the EUB 5500. Also, I read the manual of 5500 and did not find any useful thing. Let me know what should I do, if you can help me.
to rate and post a comment
Alex Mykhaylyk
re: re: re: re: re: Hitachi EUB 8500
September 09, 2015 07:19
locate the connector CH5, look at page 82, where indicated what conclusions need to connect to your mains voltage.
to rate and post a comment
Ali Deylami
re: re: re: re: re: re: Hitachi EUB 8500
September 09, 2015 11:22
I have found a table for CN5 at page 82 as follows:
Connector
No.
Input voltage Connection
100V /220V A1-B2, A2-B1 short
CN5 110V /230V A1-A3, A2-B1 short
120V /240V A1-B3, A2-B1 short
but, I really don't know what should I do with this. The connection of eub 8500 is different from 5500 and I have already connected a 110 volt hitachi 8500 to 220 volt and now it does not work.
I would be greatly tankful, if you clearly tell what exactly I should check.
to rate and post a comment
Alex Mykhaylyk
re: re: re: re: re: re: re: Hitachi EUB 8500
September 11, 2015 07:03
Remove the rear wall, the bottom center of the window, where the wires are connected. There is a connector CH5.
to rate and post a comment
Ali Deylami
re: re: re: re: re: re: re: re: Hitachi EUB 8500
September 13, 2015 10:37
I removed the rear window and found the CN5. What should I do next?
to rate and post a comment
Alex Mykhaylyk
re: re: re: re: re: re: re: re: re: Hitachi EUB 8500
September 14, 2015 08:31
What is the voltage in your circuit?
to rate and post a comment
Claude Coelo
re: re: re: re: re: re: Hitachi EUB 8500
November 20, 2015 04:52
I would like to load the EUB 8500 software to the disk drive but i do not have the CDROM software . Can someone provide me the software CDROM ?