HITACHI W950 SR "OFFSET ERROR"

Alexander Tecson

HITACHI W950 SR "OFFSET ERROR"

September 15, 2009 10:09

Good day! I encountered “OFFSET ERROR” during aircal and actual patient test,. But no error during warm up. The ct scan model is HITACHI W950 SR. Anybody can help me with this problem.. Thanks in advance.

Log inor Register

to rate and post a comment

(28)
(6)
(2)

Tajinder Singh Bhatia

Re: HITACHI W950 SR "OFFSET ERROR"

September 16, 2009 09:30

Hi, Tecson,

Data overflow & OF 01 Offset error are the most common but most critical error shown by all Hitachi single slice systems. Anyways, Check & clean the signal brushes, check error LED status at TXR2 and TXS2.

For further help contact me at mriserv@yahoo.com

T S bhatia
www.mriservices.co.in

Log inor Register

to rate and post a comment

Alexander Tecson

Re: HITACHI W950 SR "OFFSET ERROR"

September 17, 2009 11:13

Hi! T S bhatia,

Thanks for the reply,.I found out the supply of TXR2 is faulty (5v 5A) thats why all leds of that board is off. Thanks a lot T.S.Bhatia.

Log inor Register

to rate and post a comment