Hologic 1000 plus error message calibration time out

(1)

Robert Kepner

Hologic 1000 plus error message calibration time out

June 24, 2008 02:56

I am getting an error message calibration time out when I try to scan the phantom. Replaced the D/A converter and PMT. Still same message. Any ideas??

Log inor Register

to rate and post a comment

Rafael Cruz

Re: Hologic 1000 plus error message calibration time out

June 27, 2008 08:45

Check the speed.

Log inor Register

to rate and post a comment

(139)
(11)
(2)

ALEXANDER PILAFTSIS

Re: Hologic 1000 plus error message calibration time out

September 09, 2009 09:35

Dear Rafael,

I have a problem with a QDR 1000.

After we pushed F10 in order to start scanning there was the blinking sign of sichronizing but it did not go further.
I got in to survey.exe and I received NO FILTER WEEL PICKOFF.
I start again the software in order to make a scan and this time after the blinking sichronizing I received NO FILTER WEEL PICKOFF DETECTED- check motor.
I checked the motor and it was rotating.
What my be the problem?
Do I have to make the sychronization calibration or to replace part of the Filter Weel Pickoff assembly?

Thank you in advance.

Alexander Pilaftsis

Log inor Register

to rate and post a comment

(80)
(23)

Fysares Delali

Re: Hologic 1000 plus error message calibration time out

December 25, 2009 09:00

I have same problem now, how can you solve this problem???

Log inor Register

to rate and post a comment