Hi dear ,
I have a trouble with hologic delphy, I test system and at moment procedure calibration, the system show me on screen, THE X-RAY CONTROLLER IS NOT GENERATING A/C LINE INTERRUPTS
I check all connections and A/C line is connected
Some help me with service manual or procedure for fixet this trouble
thanks

Log inor Register

to rate and post a comment

(217)
(90)

MRAdmin

re: Hologic Delphy trouble

October 13, 2023 03:52

Check the IRQ line in your bios

2 people liked this
Log inor Register

to rate and post a comment

(27)

(1)

AKIS PAPAS

re: Hologic Delphy trouble

October 17, 2023 10:42

Hi, check at the working position under the power switches ,the X-Ray thermal fuse that has popped out!