Hi friends
I'm working bone densitometer delphy
The system have a lot trouble, some time the calibration pass fine but some time not pass and show error that phantom out position, this happen because de arm not move for scan phantom
Y calibrate AY movement and this action no problem, now show me other error XRAY FIRING ORDER IS OUT SEQUENCE. STATE MACHINE ERROR IN TASK 0 INDEX 278
the calibration AY appear on graph mode not values like service manual, appear other character's not values numbers or letter
I think so that software is damaged
Some help me with this
Thanks

Log inor Register

to rate and post a comment

(27)

(1)

AKIS PAPAS

re: Hologic Delphy trouble

October 17, 2023 10:42

Hi,don't mess with the software, you have to fix the filter drum first. Do replace the timing belts and check with the pin if timing will be OK after their replacement.Also, remove the 3 allen screws that hold the base of this mechanism, put it aside and check inside the precolimator cup for oil leak (rempve the 2 outer screws of the round lead colimator). If you see oil, make sure that you will retighten all 8 screws of the cup using the correct torque qnd sequence.

Log inor Register

to rate and post a comment

(148)
(25)

Germni

re: re: Hologic Delphy trouble

October 19, 2023 10:19

Thank you, I detect fail battery bios then replace battery y configuration serial and paralelo port
Now solve all trouble

Log inor Register

to rate and post a comment

(217)
(90)

MRAdmin

re: Hologic Delphy trouble

October 18, 2023 10:18

is this a new installation of a used system? or it is an old installation.

you have many problems, and the system need calibrations and checks

Log inor Register

to rate and post a comment

(148)
(25)

Germni

re: re: Hologic Delphy trouble

October 19, 2023 10:19

Thank you

Log inor Register