Hologic Discovery QC Failed

Tags: Bone Densitometer, HOLOGIC

(1)

Jaime Norales

Hologic Discovery QC Failed

December 11, 2019 10:22

My Daily QC failed .
Specifically it is referring to the Flatness test.
The error is:
The differences between the boxes 17.846154 for high air and 17.410256 for low air.

Can anyone tell me how to resolve this? We did some remodeling and the table was oved around quite a bit.

Log inor Register

to rate and post a comment

(217)
(90)

MRAdmin

re: Hologic Discovery QC Failed

December 16, 2019 10:54

The table can affect the flatness test, I can't understand what happened to the table?

Log inor Register

to rate and post a comment

(31)
(5)

Shawn McNeir

re: Hologic Discovery QC Failed

December 20, 2019 09:49

We did a short move from one room to another 20 feet down the hall and they had to perform calibrations because the table was moved. Possibly, this is the problem you are facing if the table was jostled about.