My Daily QC failed .
Specifically it is referring to the Flatness test.
The error is:
The differences between the boxes 17.846154 for high air and 17.410256 for low air.
Can anyone tell me how to resolve this? We did some remodeling and the table was oved around quite a bit.
We did a short move from one room to another 20 feet down the hall and they had to perform calibrations because the table was moved. Possibly, this is the problem you are facing if the table was jostled about.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Jaime Norales
Hologic Discovery QC Failed
December 11, 2019 10:22
My Daily QC failed .
Specifically it is referring to the Flatness test.
The error is:
The differences between the boxes 17.846154 for high air and 17.410256 for low air.
Can anyone tell me how to resolve this? We did some remodeling and the table was oved around quite a bit.
to rate and post a comment
MRAdmin
re: Hologic Discovery QC Failed
December 16, 2019 10:54
The table can affect the flatness test, I can't understand what happened to the table?
to rate and post a comment
Shawn McNeir
re: Hologic Discovery QC Failed
December 20, 2019 09:49
We did a short move from one room to another 20 feet down the hall and they had to perform calibrations because the table was moved. Possibly, this is the problem you are facing if the table was jostled about.