Hi everybody, we have the following problem with a Hologic APEX Software 12.7.3.1 when trying to execute whole body scan on hologic discovery W model:
Scan ending due to error condition.
Details:
Unable to position device within specified tolerance. State machine error in Task 0 Index 188.
Could anyone help me with this?
Thanks in advance.
Many index errors originate in the filter drum/aperture assembly. I would try the aperture alignment in the service mode. Don't forget to verify the drop pin check. These checks/procedures are found in section 4 of the service manual. Also make sure none of the cabling is on top of the source where the aperture can hit them.
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Bassam Tayar
Hologic error/Index 188
April 04, 2015 10:13
Hi everybody, we have the following problem with a Hologic APEX Software 12.7.3.1 when trying to execute whole body scan on hologic discovery W model:
Scan ending due to error condition.
Details:
Unable to position device within specified tolerance. State machine error in Task 0 Index 188.
Could anyone help me with this?
Thanks in advance.
to rate and post a comment
Desmond Johnson
re: Hologic error/Index 188
April 06, 2015 08:00
Many index errors originate in the filter drum/aperture assembly. I would try the aperture alignment in the service mode. Don't forget to verify the drop pin check. These checks/procedures are found in section 4 of the service manual. Also make sure none of the cabling is on top of the source where the aperture can hit them.
to rate and post a comment
Desmond Johnson
re: Hologic error/Index 188
April 07, 2015 07:48
The error could be coming from the table positioning pots. I would also check the table positioning pots in the service mode for linearity.
to rate and post a comment
Bassam Tayar
re: re: Hologic error/Index 188
April 07, 2015 04:34
Thank you very much for your suggestions. We will try this.
to rate and post a comment
Lenny Place
re: Hologic error/Index 188
April 08, 2015 10:04
ensure the tension for the belts are correct and they are loose the belt will sag slightly if loose.
to rate and post a comment
MRAdmin
re: Hologic error/Index 188
April 13, 2015 10:06
Hi,
Try to check the mechanical adjustment for the belt tension and the motor calibration.
if you need any help contact me: mri.service@yahoo.com