Dear Friends I had to remove the carm of the hologic explorer qdr unit to move to another room in the hosp. After reinstallation I get error message while QC procedure right after test image completed (39 second) "test failed - set SD exceeded. Phantom out of position- absolute max or min limit exceeded- an error has occurred in a csfstream" please advice. Thank you!
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
Emrah Kambak
Hologic Exporer Problem
January 28, 2018 11:25
Dear Friends
I had to remove the carm of the hologic explorer qdr unit to move to another room in the hosp. After reinstallation I get error message while QC procedure right after test image completed (39 second) "test failed - set SD exceeded. Phantom out of position- absolute max or min limit exceeded- an error has occurred in a csfstream" please advice. Thank you!
to rate and post a comment
MRAdmin
re: Hologic Exporer Problem
February 05, 2018 09:36
check the x-ray alignment, from the error it seems that the x-ray reaching the detector is low.
to rate and post a comment
Emrah Kambak
re:re: Hologic Exporer Problem
February 05, 2018 09:51
Thank you.