mehmet Oncel
Hologic Lorad M-IV restart error
October 31, 2016 10:52
Help me
Error video : https://www.youtube.com/watch?v=CQrRfujIh0s&feature=youtu.be
to rate and post a comment
Lenny Place
re: Hologic Lorad M-IV restart error
November 01, 2016 10:57
Check you power supply in the bottom of the console?
to rate and post a comment
mehmet Oncel
re: re: Hologic Lorad M-IV restart error
November 09, 2016 10:24
thank you
to rate and post a comment
Lenny Place
re: re: re: Hologic Lorad M-IV restart error
November 10, 2016 10:33
You are welcome
to rate and post a comment
Ygor Pereira
re: Hologic Lorad M-IV restart error
November 01, 2016 10:57
Hello Oncel, probably you need adjust the power supply on console.
If you want please, add me on whatsapp or Skype.
My e-mail is ygor_pereira@yahoo.com.br
to rate and post a comment
mehmet Oncel
re: re: Hologic Lorad M-IV restart error
November 09, 2016 10:24
thank you
to rate and post a comment
antoninus andres
re: re: re: Hologic Lorad M-IV restart error
March 13, 2017 10:28
hi, did you found out what causes the error? and how it will be fix? i have same lorad condition with you.. thanks
to rate and post a comment
antoninus andres
re: re: re: Hologic Lorad M-IV restart error
March 13, 2017 10:28
hi, did you found out what causes the error? and how it will be fix? i have same lorad condition with you.. thanks
to rate and post a comment
antoninus andres
re: Hologic Lorad M-IV restart error
March 06, 2017 09:03
hi, did you found out what causes the error? and how it will be fix? i have same lorad condition with you.. thanks
to rate and post a comment
Lenny Place
re: re: Hologic Lorad M-IV restart error
March 06, 2017 04:52
Check the +5vdc in console it is know to measure a little low and causes the console to reboot
to rate and post a comment
Rainner Lopez
re: re: re: Hologic Lorad M-IV restart error
June 05, 2018 06:21
hi lenny, i'm having the same trouble, i already check the power supply and it's fine, i don't know if the problem is in the host microprocessor board or in the operator console microprocessor board
to rate and post a comment
Lenny Place
re: re: re: re: Hologic Lorad M-IV restart error
June 06, 2018 02:58
Last time I had this problem and it was not the power supply it was the console microprocessor.
to rate and post a comment
RainnerLopez
re: re: re: re: re: Hologic Lorad M-IV restart error
June 08, 2018 09:34
is there a chance that do you have the Data inside the Console microprocessor memories??
to rate and post a comment
RainnerLopez
re: re: re: re: re: Hologic Lorad M-IV restart error
June 08, 2018 09:37
is there any chance you have the data of the console microprocessor memories?? or the host microprocessor memories???
to rate and post a comment
Lenny Place
re: re: re: re: re: re: Hologic Lorad M-IV restart error
June 10, 2018 09:15
I believe they were 27C4096 256x16 I am sending you the schematics they will show you the IC part Numbers
to rate and post a comment
ADRIAN MAHMUD
re: re: re: re: re: re: re: Hologic Lorad M-IV restart error
April 12, 2021 10:01
Hello, I have the same problem and in diagnosis it gives me "Memory U19 error". The source measures well. Is it enough to change the memory battery or do I also have to reinstall the software? Thank you
to rate and post a comment
RainnerLopez
re: Hologic Lorad M-IV restart error
June 08, 2018 09:37
hi friend, i'm having the same problem trouble, how do you fix that problem?? i already checked the power supply and it's fine, if you can help let me know
Rainner Lopez,Venezuela
Rainner.Lopez@hotmail.com
to rate and post a comment
Sergiy Shurayev
re: re: Hologic Lorad M-IV restart error
June 13, 2018 07:44
Two times I fixed such problems by soldering console power supply. Bad contacts.
to rate and post a comment