yikwang

Hologic Mammography System (Loard Affinity)

July 18, 2014 10:05

I'm troubleshooting the mammography system and found out error 79 Tubehead_status_error and also red led indicator on tubehead control board (LED=D7) motor over current as manual show.
Any suggestion for solving this problem?

1 person liked this
Log inor Register

to rate and post a comment

(11)
(5)

yikwang

re: Hologic Mammography System (Loard Affinity)

July 18, 2014 05:58

Anyone have the schematic diagram for lorad affinity
Please send me
hong_carlos@hotmail.com

Log inor Register

to rate and post a comment

(99)
(11)
(6)

Yojin Ramones

re: Hologic Mammography System (Loard Affinity)

July 19, 2014 09:24

Check cable on the collimator, clean verify motor mirrow, check that all be clean, because the problem is there in the collimator.

Log inor Register

to rate and post a comment