EFREN FRANCO

Hologic QDR 4500 bone densitometer problem

February 11, 2017 03:36

Hello: We have an Hologic bone densitometer and is showing the next messages "The xray controller is not generating A/C lines interrupts" and "check the instrument power switch and the x ray enable key". We have checked the hardware but still can´t find the problem origin. Please let us know posible causes about this. Thanks

Log inor Register

to rate and post a comment

(8)
(1)

Henry Collaguazo

re: Hologic QDR 4500 bone densitometer problem

February 12, 2017 12:03

Efren
Check the software configuration.

Log inor Register

to rate and post a comment

(75)
(17)
(7)

Ben Williams

re: Hologic QDR 4500 bone densitometer problem

February 13, 2017 10:34

You may have a bad XRC. If you have checked the e-stop and your electrical connections under the arm near the foot of the table you likely have a bad XRC. If you need additional assistance feel free to email. ben@dexasolutions.com

Log inor Register

to rate and post a comment

(217)
(90)

MRAdmin

re: Hologic QDR 4500 bone densitometer problem

February 19, 2017 10:43

what is the SW version that you have? is it DOS, or windows and which version? check the Hologic card is installed and configured also check the IRQ for this card.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS Mx-8000 IDT/Brilliance 16 Slice CT Scanner -----Worldwide leader!!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)