HOLOGIC QDR 4500 PROBLEM

(19)
(1)
(2)

Eduardo Firmiano

HOLOGIC QDR 4500 PROBLEM

September 21, 2010 03:34

Dear All,
I´m having problem when I try to perform "BIGFLAT". The system starts the procedure, but stops after move to right and the to left. After some seconds, the message "time out". I checked the error log and it says:
"E_STATE_MACH_RUN_ERR_S 30c101d 09/20/10 17:34:32 SCANSQ Version 1.38
State machine error: Task 0 Index 320.
Task 0 Index 320"
Could anyone help me?
Thanks

Log inor Register

to rate and post a comment

(17)

(1)

Jailyn Primus

Re: HOLOGIC QDR 4500 PROBLEM

September 30, 2010 09:31

Check your calibration data, sounds like you are missing the flattening files or they are corrupted.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

Re: HOLOGIC QDR 4500 PROBLEM

October 03, 2010 08:46

Hello Eduardo:

probably you are missing encoder pulses...

there is no problem when run mechanical/positional test??

Log inor Register

to rate and post a comment

(19)
(1)
(2)

Eduardo Firmiano

Re: HOLOGIC QDR 4500 PROBLEM

October 13, 2010 02:58

No errors when run mechanical testes

Log inor Register

to rate and post a comment

(19)
(1)
(2)

Eduardo Firmiano

Re: HOLOGIC QDR 4500 PROBLEM

October 13, 2010 02:58

More details:
I reenstaled the software. Also I tried to perform the aperture calibration and it does not complete.
the test starts, and stops after a few seconds and give the message E_INVALID_SIZE

Log inor Register

to rate and post a comment

(19)
(1)
(2)

Eduardo Firmiano

Re: HOLOGIC QDR 4500 PROBLEM

October 13, 2010 02:58

More details:
I reenstaled the software. Also I tried to perform the aperture calibration and it does not complete.
the test starts, and stops after a few seconds and give the message E_INVALID_SIZE

Log inor Register

to rate and post a comment

(15)
(3)

Henry Spong

Re: HOLOGIC QDR 4500 PROBLEM

January 02, 2011 11:43

Hi Eduardo Firmiano
Have a probleem w/t a unit and Im looking for the SOFTWARE for the Hologic 4500C if you can give mi some help please send email and prices
regards
henry

Log inor Register

to rate and post a comment

(19)
(1)
(2)

Eduardo Firmiano

Re: HOLOGIC QDR 4500 PROBLEM

February 04, 2011 05:26

Hi Henry
please, contact me at eduardo@sigmamedical.com.br
tks

Log inor Register

to rate and post a comment

(10)
(1)

carlos yanes

Re: HOLOGIC QDR 4500 PROBLEM

February 09, 2011 02:08

Subject: Re: Hologic QDR 4500 Service manual or deinstall instructions
Comment: Mi friend
I am from Venezuela
I need the instalacion manual for the software for the hologiq qdr-4500
I dont have rhe disk drive and I need to instal the software from scrash.
saludos
carlos

Log inor Register

to rate and post a comment

(13)
(4)

William Castillo

Re: HOLOGIC QDR 4500 PROBLEM

December 11, 2011 11:01

hi guy, what its your specific problem?
I can help you. Send me and email.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)