Hello: I have some problem whit a Hologic QDR1500 bone densitometer.
The QC its out of range and image its very pixeled.Software run
ok.Anybody knows this equipment.
Thanks in advance.

Log inor Register

to rate and post a comment

(21)

Patricia Cordero

Re: Hologic QDR1500

May 08, 2004 10:05

Hi Marcelo.

I guess could be a problem with the detector, there is a voltage supply in the arm of the equipment, its a dc-dc converter and the Vout is a value of several hundreds of volts DC. check this.


Enrique Rios A:
ayremedico@yahoo.com.mx

Log inor Register

to rate and post a comment

Marcelo Somavilla

Re: Hologic QDR1500

May 11, 2004 08:04

Enrique, thank you for your help.

Marcelo Somavilla

Log inor Register

to rate and post a comment

(5)

Diego Borromeo

Re: Hologic QDR1500

June 06, 2006 09:33

Hi, does anyone know the upgrade to solve the Y2K problem, with the 7.20 software version for a QDR1500?, Thanks in advance

Log inor Register

to rate and post a comment

(70)
(11)

zain abideen

re: Re: Hologic QDR1500

March 14, 2013 03:15

I HAVE QDR 2000 HOLOGIC DEXA PLZ HELLP ME