I have a problem that during the QC the error (scan ending due to error condition; Device time out waiting for the requested action. Locus State Machine. State machine error in Taks 0 index 272
Dear Mr. Gomes
I have installed a Hologic QDR4500 but I have not technical manuals and Schematics and need it. Do you have Schematics and service manuals? can you send me?
Best Regards
Kazem
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
Henrique Gomes
HOLOGIC QDR4500 elite Problem
June 15, 2011 06:47
I have a problem that during the QC the error (scan ending due to error condition; Device time out waiting for the requested action. Locus State Machine. State machine error in Taks 0 index 272
Any help will be greatly appreciated.
Henrique.
to rate and post a comment
Kazem Saremi
Re: HOLOGIC QDR4500 elite Problem
June 19, 2011 12:37
Dear Mr. Gomes
I have installed a Hologic QDR4500 but I have not technical manuals and Schematics and need it. Do you have Schematics and service manuals? can you send me?
Best Regards
Kazem
to rate and post a comment
Juan de la Cruz Martinez
Re: HOLOGIC QDR4500 elite Problem
June 20, 2011 05:33
Can you enter in service and ak the table for movement test
to rate and post a comment
Victor96
re: HOLOGIC QDR4500 elite Problem
September 15, 2016 09:40
Might someone help with the manual of service?
to rate and post a comment
MRAdmin
re: HOLOGIC QDR4500 elite Problem
September 18, 2016 12:30
Hi,
Check the Filter Drum ASSY, it could be that the belt is over tightened or bad motor.
to rate and post a comment