Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

HORIZONTAL COUCH MOVEMENT IMPOSSIBLE

Tags: CT, PHILIPS

(35)
(8)

manofxter

HORIZONTAL COUCH MOVEMENT IMPOSSIBLE

March 12, 2025 10:17

Dear Engineers,
I had a challenge with couch horizontal movement on Philips Brilliance Big Bore CT. The movement freezes with mechanical sound, reason I replace the Horizontal motor but unfortunately, the movement is not possible again.
Kindly advise

Log inor Register

to rate and post a comment

(172)
(40)
(3)

Dmitry Dederer

re: HORIZONTAL COUCH MOVEMENT IMPOSSIBLE

March 13, 2025 11:25

1. Try to calibrate the couch
2. Check teflon bullscrew
3. Check horizontal encoder
4. Check horizontal rails

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

re: HORIZONTAL COUCH MOVEMENT IMPOSSIBLE

March 13, 2025 11:25

If the cough is moving until a mechanical sound stops it, the motor is not the cause. It's a big bore, probably a big and heavy guy twisting some part of the mechanical.
First, try to move it freely by hand pushing the pedals to release it. If it stops again, you need to disarm until you can see the cause.
Best regards

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)