Hello:
I'm having a problem with an HP Image Point HX ultrasound system and message "System
Self-Test in Progress, please wait" appears on the screen and
stays forever. On PROC/GRAPH Board appears the Error EF9A3. what sign this error? and how resolve issue?
I will appreciatte any help.
Thanks.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Zoran Kolevski
HP image point error EF9A3
February 07, 2012 02:53
Hello:
I'm having a problem with an HP Image Point HX ultrasound system and message "System
Self-Test in Progress, please wait" appears on the screen and
stays forever. On PROC/GRAPH Board appears the Error EF9A3. what sign this error? and how resolve issue?
I will appreciatte any help.
Thanks.
to rate and post a comment
Brian Seltzer
Re: HP image point error EF9A3
February 07, 2012 05:15
sounds like the processor/graphics board or the power supply
to rate and post a comment
Zoran Kolevski
Re: HP image point error EF9A3
February 08, 2012 09:58
Thanks for your reply.
colleagues and I, have found that the processor board on the measurement point has 4.8VDC instead of 5VDC.
Probably error is because the low voltage..
Please advise.
to rate and post a comment
Saqueeb Usmani
Re: HP image point error EF9A3
February 08, 2012 08:49
this is cpu problem so change the board