Hello:
I am having a problem with an HP ImagePoint HX ultrasound. When performing transcranial studies sometimes there is a failure when using the Doppler. I get the message "An error has ocurred" EF512 - 00009060. I will appreciate any help you can provide. Thank you.
Regards,
Elias Valdes

Log inor Register

to rate and post a comment

Oral Unal

Re: HP ImagePoint

October 04, 2005 10:41

power supply out of regulation. Adjust gain pot. Check transmit power supply

Necati Unal

Log inor Register

to rate and post a comment

Jim Carr

Re: HP ImagePoint

October 04, 2005 11:55

From our systems engineer:
EF512
Scanner Fatal - Power supply regulation, Check Transmit supply. (GAIN control can cause "this error in certain setups") Does not say if TCD is involved...
Possible the xdcr could be bad...

rr

Log inor Register

to rate and post a comment

(58)
(6)

Elias Valdes

Re: HP ImagePoint

October 11, 2005 11:04

Thank you for your help.