Hello:
I am having a problem with an HP ImagePoint HX ultrasound. When performing transcranial studies sometimes there is a failure when using the Doppler. I get the message "An error has ocurred" EF512 - 00009060. I will appreciate any help you can provide. Thank you.
Regards,
Elias Valdes
From our systems engineer:
EF512
Scanner Fatal - Power supply regulation, Check Transmit supply. (GAIN control can cause "this error in certain setups") Does not say if TCD is involved...
Possible the xdcr could be bad...
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Elias Valdes
HP ImagePoint
October 03, 2005 08:22
Hello:
I am having a problem with an HP ImagePoint HX ultrasound. When performing transcranial studies sometimes there is a failure when using the Doppler. I get the message "An error has ocurred" EF512 - 00009060. I will appreciate any help you can provide. Thank you.
Regards,
Elias Valdes
to rate and post a comment
Oral Unal
Re: HP ImagePoint
October 04, 2005 10:41
power supply out of regulation. Adjust gain pot. Check transmit power supply
Necati Unal
to rate and post a comment
Jim Carr
Re: HP ImagePoint
October 04, 2005 11:55
From our systems engineer:
EF512
Scanner Fatal - Power supply regulation, Check Transmit supply. (GAIN control can cause "this error in certain setups") Does not say if TCD is involved...
Possible the xdcr could be bad...
rr
to rate and post a comment
Elias Valdes
Re: HP ImagePoint
October 11, 2005 11:04
Thank you for your help.
to rate and post a comment