HP Sonos 5500 : Problem

July 05, 2010 11:46

Hello,

I have a problem with Sonos 5500 HP system (Ver A.1). The symptoms is the system boot up normally but when it completes the boot up the echo image is white in color. I have tried swapping the entire Front End boards with another system (Ver B.0) all is working fine except for Beamformer I/O but when I tried to put the working Beamformer I/O board to the faulty system, it doesn't boot up at all. (Error msg: Front End no communication error). Please advise on the problem as urgently as possible. Thank you.

Nesa

Log inor Register

to rate and post a comment

(63)
(11)

Ronnie Aguiar

Re: HP Sonos 5500 : Problem

July 09, 2010 09:07

i have service manual, please check power supply, imageassistance@uol.com.br

Log inor Register

to rate and post a comment

Bill Decker

Re: HP Sonos 5500 : Problem

July 09, 2010 09:31

You have either software error, or defective front end boards. Have you run the Resident Self Tests to verify any errors in the Scanner? Contact Bill Decker at 858-997-9485 for assistance.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Philips Achieva 1.5T MRI Spare Parts Available Looking for Philips Achieva 1.5T MRI Parts?

We offer reliable & cost-effective spare parts for the Philips Achieva 1.5T MRI system helping you minimize downtime & keep operations smooth. Wide inventory available. Quick turnaround time. Expert support. Worldwide delivery. Contact Limor@elsmed.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)