Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

HSA CTi boot problem

Tags: CT

(53)
(9)

EFREN FRANCO

HSA CTi boot problem

July 09, 2018 08:44

We have a boot problem in a GE CTi system with octane host computer. the sequence of the startup is:
- Running power on diagnostics...
- Starting up the system...
- The next message appears: "PANIC: CPU 0: vfs_mountroot: no root found"
"Reboot started from CPU 0"
"CPU 0 rebooting"

If anyone have an idea about this fault I will appreciate any help about it. Thank you.

Log inor Register

to rate and post a comment

(74)
(15)

PIEBENG APPOLIN

re: HSA CTi boot problem

July 10, 2018 10:24

Octane CPU problem need to be replace.

Log inor Register

to rate and post a comment

(8)
(2)

Scott Anderson

re: HSA CTi boot problem

July 10, 2018 10:24

try reseating the system module, board on the left if looking at the rear. DONT touch gold pads, if that does fix get another cpu it will be the big aluminum thing that looks like a heat sink

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: HSA CTi boot problem

July 10, 2018 10:24

Reload software

Log inor Register

to rate and post a comment

(172)
(26)
(1)

Marivic Mendoza

re: HSA CTi boot problem

July 10, 2018 10:25

you have probably a defective hard drive. check the 3 hard drives. another maybe you have a corrupted software. try to reload

Log inor Register

to rate and post a comment

(172)
(26)
(1)

Marivic Mendoza

re: HSA CTi boot problem

July 10, 2018 10:25

you have probably a defective hard drive. check the 3 hard drives. another maybe you have a corrupted software. try to reload

Log inor Register

to rate and post a comment

(53)
(9)

EFREN FRANCO

re: re: HSA CTi boot problem

July 18, 2018 09:01

Thank you all for your advises, I changed the hard drive and reload software but when the system comes up, there is some problem with the tube configuration and the gantry turns off. I´ve already tried with several system state MO disks.

Log inor Register

to rate and post a comment

(172)
(26)
(1)

Marivic Mendoza

re: re: re: HSA CTi boot problem

July 19, 2018 08:16

open your gantry and check the model and some details of the tube.you must enter the right details of the tube.

Log inor Register

to rate and post a comment

(53)
(9)

EFREN FRANCO

re: re: re: re: HSA CTi boot problem

July 19, 2018 05:40

Thank you, problem solved, I was choosing a wrong tube type.