IE33 error codes

Tags: Ultrasound

Ross Corbett

IE33 error codes

June 14, 2013 01:51

Does anyone have a list of the IE33 error codes?

And if anyone knows what error code 034 is that would be helpful.
Thanks

Log inor Register

to rate and post a comment

Rick Reed

re: IE33 error codes

June 14, 2013 05:16

Code 034
Condition
Calibration Error On The Omni Motor Controller
Description
The system has detected a failure in calibration for the motor controller.
Corrective Action
Straighten the TEE probe and disconnect and reconnect transducer.
Scanning cannot begin until calibration is successful. If the problem
persists: replace the transducer or front-end controller.