John Vanderwerf

image intensifier problem

October 24, 2009 08:23

The I.I. has no output but all voltages are there, hear ticking (arcing) from bottle, lost vacuum? Any help will be much appreciated jvanderwerf@aol.com

Log inor Register

to rate and post a comment

(38)
(1)
(16)

Felix Donoghue

Re: image intensifier problem

October 24, 2009 11:58

disconnect only the white anode cable and try again, if it still arcs, then depending on the age and manufacturer it may be possible to repair.
hope this helps.

Log inor Register

to rate and post a comment

(21)
(2)
(3)

John Vanderwerf

Re: image intensifier problem

October 26, 2009 09:33

Hello Felix, I much appreciate your response, if it still arc's with the anode disconnected what might the problem be. Note this is a reinstall and it was working prior to the move. It's a newer 33cm Siemens I.I.

Thanks again for your help, John

Log inor Register

to rate and post a comment

(38)
(1)
(16)

Felix Donoghue

Re: image intensifier problem

October 28, 2009 10:47

If it still arcs then Grid 3 is gone or tube has leaked maybe damage while shifting. better replace. sorry not much help.