Hello great minds, I am writing from Nigeria. I need help on my IME100L, it's showing error 26 code, what do I need to do to resolve the issue? Please help
Error 26 indicates that the specified tube voltage is not provided.
when 6ms or more have passed since the beginning of x-ray generation, the X-ray tube tension on the anode or cathode side stops at 10 kV or less, or the tube tension falls below the specified value.
1. Check the connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity)
2. Check the terminal card of the generator (loosen the screws in T1, T2, B1, or B2)
The specified tube voltage is not supplied
Error code "26"
When 6 ms or more has elapsed after the start of X-ray generation, the X-ray tube voltage
on the anode or cathode side is stopped at 10 kV or less or the tube voltage falls below
than the specified value.
(a) Check connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity).
(b) Check the terminal board of the generator (looseness of the screws at T1, T2, B1, or
B2).
(c) Readjust the IME-CONT PWB.
Two months ago I'm in Nigeria
Error code "26"
When 6 ms or more has elapsed after the start of X-ray generation, the X-ray tube voltage
on the anode or cathode side is stopped at 10 kV or less or the tube voltage falls below
than the specified value.
(a) Check connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity).
(b) Check the terminal board of the generator (looseness of the screws at T1, T2, B1, or
B2).
(c) Readjust the IME-CONT PWB.
Two months ago I'm in Nigeria
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Akpama Isokon
IME100L error 26
March 18, 2026 09:56
Hello great minds, I am writing from Nigeria. I need help on my IME100L, it's showing error 26 code, what do I need to do to resolve the issue? Please help
to rate and post a comment
CRAIG WAMBURA
re: IME100L error 26
March 19, 2026 02:20
Error 26 indicates that the specified tube voltage is not provided.
when 6ms or more have passed since the beginning of x-ray generation, the X-ray tube tension on the anode or cathode side stops at 10 kV or less, or the tube tension falls below the specified value.
1. Check the connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity)
2. Check the terminal card of the generator (loosen the screws in T1, T2, B1, or B2)
3. Adjust the IME-CONT PWB
Contact me for more details
to rate and post a comment
Khawar Mukhtar
re: IME100L error 26
March 19, 2026 02:20
The specified tube voltage is not supplied
Error code "26"
When 6 ms or more has elapsed after the start of X-ray generation, the X-ray tube voltage
on the anode or cathode side is stopped at 10 kV or less or the tube voltage falls below
than the specified value.
(a) Check connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity).
(b) Check the terminal board of the generator (looseness of the screws at T1, T2, B1, or
B2).
(c) Readjust the IME-CONT PWB.
Two months ago I'm in Nigeria
to rate and post a comment
Khawar Mukhtar
re: IME100L error 26
March 19, 2026 02:20
Error code "26"
When 6 ms or more has elapsed after the start of X-ray generation, the X-ray tube voltage
on the anode or cathode side is stopped at 10 kV or less or the tube voltage falls below
than the specified value.
(a) Check connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity).
(b) Check the terminal board of the generator (looseness of the screws at T1, T2, B1, or
B2).
(c) Readjust the IME-CONT PWB.
Two months ago I'm in Nigeria
to rate and post a comment