IME100L error 26

Tags: Other, TOSHIBA

(5)
(1)

Akpama Isokon

IME100L error 26

March 18, 2026 09:56

Hello great minds, I am writing from Nigeria. I need help on my IME100L, it's showing error 26 code, what do I need to do to resolve the issue? Please help

1 person liked this
Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: IME100L error 26

March 19, 2026 02:20

Error 26 indicates that the specified tube voltage is not provided.

when 6ms or more have passed since the beginning of x-ray generation, the X-ray tube tension on the anode or cathode side stops at 10 kV or less, or the tube tension falls below the specified value.

1. Check the connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity)

2. Check the terminal card of the generator (loosen the screws in T1, T2, B1, or B2)

3. Adjust the IME-CONT PWB

Contact me for more details

Log inor Register

to rate and post a comment

(2)

Khawar Mukhtar

re: IME100L error 26

March 19, 2026 02:20

The specified tube voltage is not supplied
Error code "26"
When 6 ms or more has elapsed after the start of X-ray generation, the X-ray tube voltage
on the anode or cathode side is stopped at 10 kV or less or the tube voltage falls below
than the specified value.
(a) Check connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity).
(b) Check the terminal board of the generator (looseness of the screws at T1, T2, B1, or
B2).
(c) Readjust the IME-CONT PWB.
Two months ago I'm in Nigeria

Log inor Register

to rate and post a comment

(2)

Khawar Mukhtar

re: IME100L error 26

March 19, 2026 02:20

Error code "26"
When 6 ms or more has elapsed after the start of X-ray generation, the X-ray tube voltage
on the anode or cathode side is stopped at 10 kV or less or the tube voltage falls below
than the specified value.
(a) Check connections (disconnection of CNN INV1, HPG2, 702, 703; discontinuity).
(b) Check the terminal board of the generator (looseness of the screws at T1, T2, B1, or
B2).
(c) Readjust the IME-CONT PWB.
Two months ago I'm in Nigeria

Log inor Register

to rate and post a comment