Sebastien Dupont
Imed Gemini PC-4
October 24, 2002 12:35
Can someone know how to access the configuration mode on start up ? I hold the Panel Lock on power up but it doesn't work !!!
to rate and post a comment
Philip Stuemke
Re: Imed Gemini PC-4
October 24, 2002 01:59
Once the pump is started, before you press any other keys, press the options
key. That will let you into the configuration mode. From there, just use
the arrow and number keys to scroll to the option you want.
Philip Stuemke
NOVATEK Medical
to rate and post a comment
Sebastien Dupont
Re: Re: Imed Gemini PC-4
October 24, 2002 03:35
Thank you Philip but it still doesn't work ! I always have 'Error number 704' on start up and I can do nothing. According to the operator's manual, it is a software problem...
to rate and post a comment
Philip Stuemke
Re: Re: Re: Imed Gemini PC-4
October 24, 2002 03:57
704 means your internal nicad battery is depleted or completely dead. Have you tried charging the unit? Unless your battery is completely dead, plugging the unit in should charge both the nicad battery and the main unit battery. We deal with these units all day long. If you have any more questions feel free to give me a call.
Philip Stuemke
NOVATEK Medical
618-483-9198
to rate and post a comment
Sebastien Dupont
Re: Re: Re: Re: Imed Gemini PC-4
October 24, 2002 04:28
Thanks again, I will recharge the batteries tonigh. Is it the same pattern with the PC-2TX ???
to rate and post a comment
Philip Stuemke
Re: Re: Re: Re: Re: Imed Gemini PC-4
October 24, 2002 05:52
Yes, same on PC2TX. If you need anything else you can email me at prstuemke@novatekmedical.com
That would be faster and less hassle than going through the dotmed reply.
Philip