Innova 2000 table and c- arm problem

Tags: Radiology - R/F

(59)
(2)

Abu-Yusuf

Innova 2000 table and c- arm problem

July 15, 2014 10:13

Hi all, i have problem with innova 2000 OMEGA-IV Compact Table. Its not moving i have checked all fuses and switches they are ok. What can be the problem?
Thanks in advance

Log inor Register

to rate and post a comment

(54)
(9)

Mina halim

re: Innova 2000 table and c- arm problem

July 16, 2014 09:15

r u sure that it has fed by the supply?

Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: Innova 2000 table and c- arm problem

July 16, 2014 09:15

Is not moving Horizontal or Vertical? Have you check the safeties ?

Log inor Register

to rate and post a comment