Can the Philips Intera MRI magnet(F2000) be rumped up using the power supply of the Philis Gyroscan NT?
Thank you

1 person liked this
Log inor Register

to rate and post a comment

(26)
(2)
(1)

Nenad Kostic

re: Install Philips Intera 1.5T - MRI - Rump up

September 18, 2015 07:34

Yes. It's possible. If you change cables. And leads. And if you are able to modify power supply to read and operate with MIB.
Probably its cheaper and much more safety for all to rent properly PS.