When using remote, image, run or snapshot functions work fine. When trying to use STEP to cycle thru one image at a time, system cycles thru several images without stopping. We tried a new remote and got the same symtoms. We replaced the VCCOM 2 board and got the same symtoms. We have eliminated the IR Receiver. Any suggestions as to the root of the problem would be greatly appreciated.
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Jamie Reid
Integris H3000 VISUB Problem
January 24, 2007 10:26
When using remote, image, run or snapshot functions work fine. When trying to use STEP to cycle thru one image at a time, system cycles thru several images without stopping. We tried a new remote and got the same symtoms. We replaced the VCCOM 2 board and got the same symtoms. We have eliminated the IR Receiver. Any suggestions as to the root of the problem would be greatly appreciated.
Thank you.
to rate and post a comment
Jorge Cid Alvarez
Re: Integris H3000 VISUB Problem
April 01, 2007 02:13
Dear Sirs,
I have about 4 years experience on VISUB.
it looks as software problem
may the best were reinstall the options
JORGE CID ALVAREZ
electronic engineer.