iU22 error 020

Tags: Ultrasound

(22)
(2)
(3)

li ming

iU22 error 020

October 07, 2013 08:57

I have a A cart iU22,with error 020,and I can see the DSC be lost,if replug cables and DSC,
then sometimes working fine for weeks,or a month,then 020 arrear again.can anyone help me to locat the problem?

Thanks

Sam

Log inor Register

to rate and post a comment

(22)
(12)
(72)

Volodymyr Markevych

re: iU22 error 020

October 07, 2013 07:24

Error 020 means hardware mismatch:
Description: The installed hardware configuration is not an allowed configuration for the currently installed software.

Corrective Action
Either use the error log browser or review the 'Expected versus Actual' list
in System Config to identify which boards have the wrong configuration and
replace as needed. Otherwise, if there is new version of host software that
supports the current hardware configuration, install host software.

This error will appear if DSC is missing.
It can be DSC, cables or software.

2 people liked this
Log inor Register

to rate and post a comment

(2)

Mehmet Akif Demirayak

re: re: iU22 error 020

February 28, 2014 02:00

Hi
I am writing from Turkey.
I need your help with something.
IE33 my phılıps software DVDs have available. but during the installation, and the installation does not provide diagnostic err. Diagnostic looks err code 602.
Can you help me about the installation procedure?
thanks

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)