Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

IU22 ERROR CODE 606

(70)
(3)

Garald jean louis

IU22 ERROR CODE 606

December 01, 2023 05:42

good day -
Im reporting an error 606 on an IU22 ultrasound from Philips . can anyone give me a clue to solve the problem please .
thanks

1 person liked this
Log inor Register

to rate and post a comment

YASEEN AKHTAR

re: IU22 ERROR CODE 606

December 04, 2023 10:34

Hi sir

Log inor Register

to rate and post a comment

(70)
(3)

Garald jean louis

re: re: IU22 ERROR CODE 606

December 05, 2023 11:39

yes it has 1 mother board and 3 hard drives

Log inor Register

to rate and post a comment

(2)
(2)

NCFSE

re: IU22 ERROR CODE 606

December 04, 2023 10:35

Need to know what cart level.
If it has 2 computers in the back(dual motherboard), this would signify that they aren’t communicating with each other.
If only one computer UMB (3 hard drives) or EMB (1 hard drive) I would replace the HDD’s .

1 person liked this
Log inor Register

to rate and post a comment

Abtech eletronica

re: IU22 ERROR CODE 606

December 04, 2023 10:35

I have already received some with this error 606, 80% had a defect in the HD sip, and older hardware had software corruption, this will depend on the type of hardware version of this iU22, as it exists, as variations (A, B, C, D, E, F, G or R), but investigate whether there is communication with the SIP subsystem with Host, using the oscilloscope if the waveform is present. I hope I have helped from a distance it is difficult to help.

Log inor Register

to rate and post a comment

Abtech eletronica

re: IU22 ERROR CODE 606

December 04, 2023 10:35

I have already received some with this error 606, 80% had a defect in the HD sip, and older hardware had software corruption, this will depend on the type of hardware version of this iU22, as it exists, as variations (A, B, C, D, E, F, G or R), but investigate whether there is communication with the SIP subsystem with Host, using the oscilloscope if the waveform is present. I hope I have helped from a distance it is difficult to help.

1 person liked this
Log inor Register

to rate and post a comment

(30)
(7)

EDGAR SANTIAGO RIVERA ZAMUDIO

re:IU22 ERROR CODE 606

December 04, 2023 10:35

clean the ram memory cards and sockets

1 person liked this
Log inor Register

to rate and post a comment

(70)
(3)

Garald jean louis

re: re:IU22 ERROR CODE 606

January 09, 2024 09:57

just remove them and clean them could help ?

Log inor Register

to rate and post a comment

(80)
(9)

Nikolas

re: IU22 ERROR CODE 606

December 04, 2023 10:36

606
Condition
SIP Power Failed To Initialize.
Description
Platform Monitor failed to start the System Power process.
Corrective Action
Possible causes: (1) SIP is not answering on named pipe or (2) Host
physical memory problem. Troubleshoot Dolphin IPL. See the article:
'MTE: Troubleshooting the IPL (T-00046)'.
INFO-
Hello, reconnect cable between Doplphin cards, otherwise - reinstall sw.
Is it hw B with 2 motherboards?

1 person liked this
Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: IU22 ERROR CODE 606

December 04, 2023 01:22

Code 606
Condition
SIP Power Failed To Initialize.
Description
Platform Monitor failed to start the System Power process.
Corrective Action
Possible causes: (1) SIP is not answering on named pipe or (2) Host
physical memory problem. Troubleshoot Dolphin IPL. See the article:
'MTE: Troubleshooting the IPL (T-00046)'.

Log inor Register

to rate and post a comment

(70)
(3)

Garald jean louis

re: re: IU22 ERROR CODE 606

January 09, 2024 09:57

ok you mean the software on SIP could be corrupted ? it has 2 other drives on the SIP box - Can you send me a picture of that firewire cable please ? - Would also want to know the parameters to put in the Bios please .

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)