Kodak 8000c movement_detection E23

August 20, 2018 02:17

I have a Kodak 8000c that was working fine up until today when it displayed an error stating "movement_detection". The error code is E23.

Does anyone have any insight into what could be causing this or a manual that will give me help with the error code and troubleshooting?

Thanks in advance.
Aaron ~ apetrosky@tech2nite.com

Log inor Register

to rate and post a comment

(330)
(228)
(4)

Faustino Pereira

re: Kodak 8000c movement_detection E23

August 21, 2018 10:02

Could be fuses on motor board, motor wires cut, encoder wires cut due to stress during rotation, encoder defective.There's an encoder mounted on the motor shaft that sends pulses to the CPU. If pulses missing or motor not running, error appears.