I have a Kodak 850 CR. It is having an error when we insert the cassette into the CR. The error code is 000951-100014. Does anybody know this error and does anyone have a list of the Kodak CR 850 error code.

Log inor Register

to rate and post a comment

(10)
(1)

Bob Davidson

re: kodak error code 000951-100014

May 05, 2012 12:49

11014 - DUPLEX CAM: Timeout Error
Cause: The DUPLEX CAM malfunctioned.
Action: Do the checkout for the CASSETTE HANDLING AY.

Log inor Register

to rate and post a comment

(10)
(1)

Bob Davidson

re: re: kodak error code 000951-100014

May 07, 2012 10:06

oophs gave 11 not 10. No need to remove sloscan, just clean and lube the rails (not the screw!) if that is the problem it should correct it. Otherwise install new sloscan assy.

Log inor Register

to rate and post a comment

(118)
(28)

Steven McMurray

re: re: re: kodak error code 000951-100014

May 07, 2012 03:11

Thank you very much. Found the problem. It was a light leak.

Log inor Register

to rate and post a comment

(38)
(8)

Le Hai

re: kodak error code 000951-100014

May 07, 2012 09:56

dear Steven!
removal the slow scan, do clean and assambly agian!!

Log inor Register

to rate and post a comment

(118)
(28)

Steven McMurray

re: re: kodak error code 000951-100014

May 07, 2012 03:11

Thank you very much. Found the problem. It was a light leak.

Log inor Register

to rate and post a comment

(118)
(28)

Steven McMurray

re: re: kodak error code 000951-100014

May 07, 2012 03:12

Thank you very much. Found the problem. It was a light leak.