A Mishra

Kodak POC

December 07, 2012 10:08

Dear Friends,

We are sometimes getting a warning on computer monitor related top Kodak POC 140.
The detailed message is: Erase lamps are working at reduced capacity. Image erasing not assured. Call service.
Please adv solution.


2 people liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Kodak POC

December 07, 2012 05:59

replace the eraser lamps they are old

1 person liked this
Log inor Register

to rate and post a comment

(21)
(2)

P402

re: re: Kodak POC

December 07, 2012 06:43

Dear Lenny

I'm looking for a solution about POC 140. It can not scan full image after pushed cassettes in. It only scan apart of immage. Now i want to make a calibration to correct field of Scanning so i don't know how to do this. Could you please help me?. My mail:ngoc.vu@ett.vn

Thanks,

Log inor Register

to rate and post a comment

(77)
(15)
(2)

Gery Rodriguez

re: re: re: Kodak POC

December 07, 2012 07:17

In this case when scanner only gives you half of an image, then the USB Board is faulty. Re-Calibrating the machine will not solve this, not to mention that if you don't have the proper tools to calibrate the CR you are looking for trouble.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Kodak POC

December 07, 2012 07:17

this is normally caused by dirty rollers or a bad sensor

Log inor Register

to rate and post a comment

(5)
(2)

devlim

re: re: re: Kodak POC

December 07, 2012 10:53

When inserting and loading a cassette, does it indeed indicate (on bottom left of screen) the correct size of cassette? If not, it maybe one of the sensors, or dirty or slipping (back)rollers, which may be solved by cleaning or replacing roller springs.

1 person liked this
Log inor Register

to rate and post a comment

(21)
(2)

P402

re: re: re: re: Kodak POC

December 09, 2012 10:17

Please give me your mail. I will send you some pictures.

Log inor Register

to rate and post a comment

(21)
(2)

P402

re: re: re: re: Kodak POC

December 09, 2012 10:54

It can scan any cassetes and display the correct size of cassetes but images are not full.

Log inor Register

to rate and post a comment

(77)
(15)
(2)

Gery Rodriguez

re: re: re: re: re: Kodak POC

December 09, 2012 11:06

Like I said, could be a bad USB board.

1 person liked this
Log inor Register

to rate and post a comment

(21)
(2)

P402

re: re: re: re: re: re: Kodak POC

December 09, 2012 11:07

Thank for help. I will tried to check it.

Log inor Register

to rate and post a comment

(21)
(2)

P402

re: re: re: re: re: re: Kodak POC

December 09, 2012 11:07

Thank for help. I will check it.

Log inor Register

to rate and post a comment

(18)
(4)

Michael Solyom

re: re: re: Kodak POC

December 10, 2012 10:15

is the image only 15 cm wide once scanned?

Log inor Register

to rate and post a comment

(18)
(4)

Michael Solyom

re: re: re: Kodak POC

January 14, 2013 10:06

Have you checked the 15 x 30 cm dental sensor which is housed in the feed tray. When this sensor is faulty the scanner will only scan 15 cm wide image. If you unplug this sensor from sensor board the fault should go away.

Log inor Register

to rate and post a comment

(77)
(15)
(2)

Gery Rodriguez

re: Kodak POC

December 07, 2012 07:17

In this particular case, depending on the software version you simply need to reisntall the software and run a (bat) file that fixes this bug. Remember to back up you anatom settings and calib files before you do so, you will need them again. You can open the reader and test the lamps if all turn on, the it's most likely that software glitch. If you want to dig even deeper, make sure the lamps power supply is not faulty.

Log inor Register

to rate and post a comment

(11)
(2)

A Mishra

re: re: Kodak POC

December 08, 2012 11:17

Thanks Gerry. I presume your diagnosis is right as the message appears quite irregularly & when we insert the cassette it when massage is displayed, the message disappears and also the screen is erased.

Log inor Register

to rate and post a comment

(11)
(2)

A Mishra

re: re: Kodak POC

January 14, 2013 12:35

Dear Gery,
Please adv. how we can locate anatom settings & calib files on the system?

Log inor Register

to rate and post a comment

(36)
(3)

yamil sanchez

re: Kodak POC

December 07, 2012 10:53

Hi friends.
We have the manual thechnical..
yamil7399@gmail.com

Log inor Register

to rate and post a comment

(77)
(15)
(2)

Gery Rodriguez

re: re: Kodak POC

January 14, 2013 03:56

go to: My Computer, Program Files, Kodak Folder, Bin.

Log inor Register

to rate and post a comment

(18)
(4)

Michael Solyom

re: Kodak POC

December 10, 2012 10:15

Dear Sir before you do anything else please replace fuse holders for fuse F1 and F2 they are the problem. Be advised that you need to replace these with a higher current rating than original as they do not last before becoming intermittently faulty again.

Log inor Register

to rate and post a comment

(4)
(1)

Maurice Galvez

re: Kodak POC

January 12, 2013 11:23

Fuse Housing, Power Supply (the middle one), connectors or the erase lamps are bad. Start with the connectors, go for the fuses (top and second from top), if not go for the fuse housing (top two), power supply (middle one) and your last resource would be the erase lamps.

If you don't want to go through all the testing, check the connections of the erase lamps and see if there is a bank that doesn't turn on.

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)