I have a problem with a kretz Voluson 530 ultrasoundunit. The problem is that when trying to make measurements we recieve the following message: "PC calculating in progress"followed by any of these numbers 59, 136 or 139. Sometimes we getthe message "PC communication error" and sometimes the systemgoes down. Any help will be appreciated. Thanks,my email is: mate.rus@hotmail.com
Problem Error message "PC Calculating in Progress XX", no 4D, no PID entry possible, no Table reports possible: Arrow on left upper corner of screen remains always flashing( longer then 2 minutes),
Described problems tell us about the same cause, the PC part (inside lower main electronic box named GAA) which is only used for special volume rendering calculations and windows functions. Often related to PC industrial board itself , or maybe some other component of PC.
Could also be caused at external drive problem.
The basic system functions B/D/CFM..are controlled by separate Motorola CPU, and therefore no problem.
Action:
Please remove the SCSI cable on rearpanel of V530D to see the influence of PC booting.
Also an error on the dicom page could create some wait signals, so please check net key-view status- error list , and clear out all errors.
Dear Sir
Can I get a service manula or soft ware CD of Voluson 530 D from any where as one of my nuit is out of order with same problem of "PC communication error"
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
mate zigo
kretz voluson 530 problem
June 08, 2009 08:27
I have a problem with a kretz Voluson 530 ultrasoundunit. The problem is that when trying to make measurements we recieve the following message: "PC calculating in progress"followed by any of these numbers 59, 136 or 139. Sometimes we getthe message "PC communication error" and sometimes the systemgoes down. Any help will be appreciated. Thanks,my email is: mate.rus@hotmail.com
to rate and post a comment
Marko Shevchuk
Re: kretz voluson 530 problem
June 08, 2009 12:59
Problem Error message "PC Calculating in Progress XX", no 4D, no PID entry possible, no Table reports possible: Arrow on left upper corner of screen remains always flashing( longer then 2 minutes),
Described problems tell us about the same cause, the PC part (inside lower main electronic box named GAA) which is only used for special volume rendering calculations and windows functions. Often related to PC industrial board itself , or maybe some other component of PC.
Could also be caused at external drive problem.
The basic system functions B/D/CFM..are controlled by separate Motorola CPU, and therefore no problem.
Action:
Please remove the SCSI cable on rearpanel of V530D to see the influence of PC booting.
Also an error on the dicom page could create some wait signals, so please check net key-view status- error list , and clear out all errors.
to rate and post a comment
Marko Shevchuk
Re: kretz voluson 530 problem
September 13, 2009 04:53
Hi!
You solve problems with Voluson?
Pls answer me.
thanks in advance
Marko
to rate and post a comment
DR. Tanveer Hasan Zubairi
Re: kretz voluson 530 problem
September 14, 2009 02:52
Dear Sir
Can I get a service manula or soft ware CD of Voluson 530 D from any where as one of my nuit is out of order with same problem of "PC communication error"
to rate and post a comment
Murat Canpolat
Re: kretz voluson 530 problem
September 15, 2009 11:00
hello sir
İf you want, I can help you abaout Voluson 530.I have Voluson 530 software.
to rate and post a comment