Beamformer board problem. The US image is fadeing in and out slowly. The beam lights up suddenly and then fades into darkness :-) and again and again.
I managed to get another BF board, but it's not the same version. With the new board there's no fade in/out, but I have horizontal stripes across the beam, so it's not good.
So, obviously the BF is faulty. Can it be repaired? Maybe someone had a problem like this?
All tests in diagnostic passed :-)
If wanted, I can upload a video of the fault.
Thanks for help!
Best regards!
Dear Kobris,the problem is very common in BF boards from Voluson.We are doing board repairs and probe repairs and we have repaired a lot of BF.Do you have anyone with knowledge in board repair to help you?
The channel boards must be ok,we have never changed a channel board.You should look at the right corner of the board,not the side with the capacitors the other side.There is a very small three leg transistor there that you have to use a magnifing lense to see it.Very common to fail.
re: re: re: re: re: Krez Voluson 730 image problem
August 22, 2014 10:08
Right now i do not have a beamformer in the company.I will have one in a few days and i will tell you exactly which one to search for.As for replacing i always order the same one from an electronics store,but the 6 you found is not the correct ones.The element i am talking about is the size of a pinhead
I have the same problem and I replaced the BF together with all the channels, it didn't solve the problem, Have you solve the problem?
Do you have any suggestion?
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Kobris
Krez Voluson 730 image problem
August 18, 2014 04:27
Hi all,
Old Voluson 730, SW v1.07f
Beamformer board problem. The US image is fadeing in and out slowly. The beam lights up suddenly and then fades into darkness :-) and again and again.
I managed to get another BF board, but it's not the same version. With the new board there's no fade in/out, but I have horizontal stripes across the beam, so it's not good.
So, obviously the BF is faulty. Can it be repaired? Maybe someone had a problem like this?
All tests in diagnostic passed :-)
If wanted, I can upload a video of the fault.
Thanks for help!
Best regards!
to rate and post a comment
Ricardo A. Sanchez
re: Krez Voluson 730 image problem
August 19, 2014 10:10
I will check the CPP too, I have seen the problem with the imagen fading out and the problem was the CPP
to rate and post a comment
Kobris
re: re: Krez Voluson 730 image problem
August 19, 2014 05:41
Hi Ricardo, thanks for your advice, I'll check.
to rate and post a comment
EROTOKRITOS SKENTOS
re: Krez Voluson 730 image problem
August 19, 2014 10:12
Dear Kobris,the problem is very common in BF boards from Voluson.We are doing board repairs and probe repairs and we have repaired a lot of BF.Do you have anyone with knowledge in board repair to help you?
to rate and post a comment
Kobris
re: re: Krez Voluson 730 image problem
August 19, 2014 05:41
Thanks Erotokritos, can you recommend where to start checking the board? I think the channel boards (on BF) are OK...
to rate and post a comment
EROTOKRITOS SKENTOS
re: re: re: Krez Voluson 730 image problem
August 20, 2014 08:24
The channel boards must be ok,we have never changed a channel board.You should look at the right corner of the board,not the side with the capacitors the other side.There is a very small three leg transistor there that you have to use a magnifing lense to see it.Very common to fail.
to rate and post a comment
Kobris
re: re: re: re: Krez Voluson 730 image problem
August 20, 2014 03:18
could you be more specific about the location of that element please? I found 6 of them...all good...and if yout good, whatz do you use to replace it?
to rate and post a comment
EROTOKRITOS SKENTOS
re: re: re: re: re: Krez Voluson 730 image problem
August 22, 2014 10:08
Right now i do not have a beamformer in the company.I will have one in a few days and i will tell you exactly which one to search for.As for replacing i always order the same one from an electronics store,but the 6 you found is not the correct ones.The element i am talking about is the size of a pinhead
to rate and post a comment
Kobris
re: re: re: re: re: re: Krez Voluson 730 image problem
August 25, 2014 10:13
Ok, please contact me at kobris@gmail.com
to rate and post a comment
Doron Forshtat
re: re: re: re: re: re: Krez Voluson 730 image problem
February 05, 2015 12:32
I have the same problem and I replaced the BF together with all the channels, it didn't solve the problem, Have you solve the problem?
Do you have any suggestion?
to rate and post a comment
Doron Forshtat
re: re: re: re: re: re: Krez Voluson 730 image problem
February 05, 2015 12:33
Dear Erotokritos
I have the same problem
Replaced the BF and it didn't help
Have you found something else?
Thank you