I need advise on how to rectify Philip practix 100, mobile x-ray maxhine given error message lacking X-ray when taking x-ray from 50KV upward, but below 49KV setting it expose normal with x-ray. Kindly advise what I can do to repair the machine.
Thanks
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Imade Nathaniel
Lacking X-ray Philip practix 100
September 22, 2016 10:24
I need advise on how to rectify Philip practix 100, mobile x-ray maxhine given error message lacking X-ray when taking x-ray from 50KV upward, but below 49KV setting it expose normal with x-ray. Kindly advise what I can do to repair the machine.
Thanks
to rate and post a comment
ROBERTO DIAZ BEDOLLA
re: Lacking X-ray Philip practix 100
September 23, 2016 09:34
check inverter board first
to rate and post a comment
umberto sciunnach
re: Lacking X-ray Philip practix 100
September 23, 2016 09:40
You must investigate tube !!
to rate and post a comment
DMEDS
re: Lacking X-ray Philip practix 100
September 24, 2016 04:20
Dear Sir,
you have to check the inverter block contain the IGBT and the driving board,if every thing is OK you have to check x ray tube it may be gases,
i have the service manuals if you need it contact me,
magmesme@gmail.com
Regards
to rate and post a comment