Engr.Ketchup

LIFEPAK20e Error 1C13

January 03, 2017 10:10

I have a problem with my defibrillator..

it is howing 1c13 on error logs,

does anybody know some way to fix this without replacing any parts?.. thanks...

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: LIFEPAK20e Error 1C13

January 04, 2017 10:34

1c13 UI FPGA Flash memory corrupted 1. Reload the device software.
2. Replace the A05 UI PCB

Log inor Register

to rate and post a comment

(1)

Owen Liu

re: LIFEPAK20e Error 1C13

January 04, 2017 10:35

Error code: 1c13
Error Description: UI FPGA Flash memory corrupted

Corrective Action:
1. Reload the device software.
2. Replace the A05 UI PCB.

This was the document I was referencing to (Page 7-19):
http://www.frankshospitalworkshop.com/equipment/documents/defibrillators/service_manuals/Medtronic_Lifepak_20_-_Service_manual.pdf

Regards,

Owen.Liu@Datrend.com

Log inor Register

to rate and post a comment

(66)
(8)
(4)

William Clarke

re: LIFEPAK20e Error 1C13

January 04, 2017 10:35

Send me your email address and I will send the manual.
Bill Clarke
wlcbmt@aol.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)