Error message opened after os initializing finished as "The dataacq process has disabled scanning due to a hardware failure. So, gesyslog described "no device driver installed Unable to open the DIP device driver. Driver name:/dev/dip or /dev/gemsdip Open mode:read/write OS error: 2---no such file or directory" pls advice to fix this problem.
Dear SungHo,
i do not have good experiance in the LS16 but i fixed some troubls( i have 13 years experiance with Picker CTs).
first i will see if i can do X-ray without collecting data then go up step by step.
please do the following:
Service
Diagnostic
Data Acquistion
Diagnostic Data Collection DDC
1-try first to do static Xray without data acqusition.
2-if it is passed then try to do static x-ray with data Acquisition if passed
3-retry both with rotating condition (not Static)
best wishs
Borhan
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
SungHo Kang
Light speed 16ct'error message
June 14, 2011 03:24
Error message opened after os initializing finished as "The dataacq process has disabled scanning due to a hardware failure. So, gesyslog described "no device driver installed Unable to open the DIP device driver. Driver name:/dev/dip or /dev/gemsdip Open mode:read/write OS error: 2---no such file or directory" pls advice to fix this problem.
to rate and post a comment
Mariusz Figoluszka
Re: Light speed 16ct'error message
June 14, 2011 06:30
I'd go to:
1. DARC reloading
2. LFC
3. look for hardware problems (ie. hdd)
to rate and post a comment
Borhan Kalash
Re: Light speed 16ct'error message
June 15, 2011 08:17
Dear SungHo,
i do not have good experiance in the LS16 but i fixed some troubls( i have 13 years experiance with Picker CTs).
first i will see if i can do X-ray without collecting data then go up step by step.
please do the following:
Service
Diagnostic
Data Acquistion
Diagnostic Data Collection DDC
1-try first to do static Xray without data acqusition.
2-if it is passed then try to do static x-ray with data Acquisition if passed
3-retry both with rotating condition (not Static)
best wishs
Borhan