LightSpeed 16 Error codes

Tags: CT, GE

(140)
(35)

FranciscoG

LightSpeed 16 Error codes

March 30, 2016 10:45

Hello, I've this LS 16 with the following errors:

Host: ETC. Proc. Scan. Error: 182804
HW/SW error: Communication Problems with the Gantry Display
The Gantry Display is NOT Initialized Properly
(Scanning is NOT allowed if the Gantry Display is NOT working)

-- this error appears sometimes and if I turn the system off for a few hours the system works perfectly fine, when it doesn't the Gantry displays continue reseting themselves eachtime.

Host: OBC. Proc: Generator. Error: 185557
Function: XRAY GENERATION: Tube Rotor Control
Could not send HEMRC message in the allotted time.
HEMRC serial link is down. HEMRC_ISO_+12V Failure detected.

-- this error won't let me make any xray exposure, even scouts or tube warmups.

Any help is appreciated.

1 person liked this
Log inor Register

to rate and post a comment

(38)
(7)
(1)

Levan Gamtsemlidze

re: LightSpeed 16 Error codes

March 31, 2016 12:17

Hi your problem is in +12V power supply unit, change it and your problem will be solved

1 person liked this
Log inor Register

to rate and post a comment

(423)
(40)
(5)

Juan de la Cruz Martinez

re: LightSpeed 16 Error codes

March 31, 2016 12:17

Check the table power Supply

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: LightSpeed 16 Error codes

May 16, 2016 10:14

Hi Juan, I didn't have the Gantry Initialization issue anymore until now, so I checked the power supplies in the table (PS1 and PS2). These are the results, I couldn't find the output voltages for each PS in the manuals.

These are the measurements:
Power input for PS1 and PS2 is 124VAC

PS1 (ON TOP) Reference is GND
PS1- 0.00 VDC
PS1+ -24.00 VDC

PS2 (ON BOTTOM) Reference is GND
-V4 PS2 0.004 VDC
+V4 PS2 -24.03 VDC
-V3 15.01 VDC
-V2+V3 -0.002 VDC
+V2 -15.11 VDC

-V1PS2 0.006 VDC
+V1PS2 -5.12 VDC

There is another Power Supply on top of the OBC module, that has the same values as PS2.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: LightSpeed 16 Error codes

May 16, 2016 10:14

Hi Juan, so I double checked the Power supplies in the table and the one on the OBC module. I found how to do it right.

The +24, +5, +15 and -15 Vdc are present in the PS2 (Bottom Table Power supply) and the OBC PS.

And there are +24 Vdc present in PS1 (Top Table Power Supply) if I turn the system off for like a day, the problem is gone, but sometimes comes back and can not do anything, cause the Displays are resetting themselves again and again.

I checked the STC power supply as well, so +5, +3.3 and +12 Vdc are present on the Axial Drive board.

I ran out of ideas on how to solve this issue, I checked the connectors going to the table and going to the STC backplane and even disconnected them and plugged them back up, with no success. If you have any idea, let me know, please.

Thanks.

Log inor Register

to rate and post a comment

(26)
(2)

Francisco Garcia

re: re: LightSpeed 16 Error codes

May 16, 2016 10:19

So I checked the Power supplies in the table and the one on the OBC module.

The +24, +5, +15 and -15 Vdc are present in the PS2 (Bottom Table Power supply) and the OBC PS.

And there are +24 Vdc present in PS1 (Top Table Power Supply) if I turn the system off for like a day, the problem is gone, but sometimes comes back and can not do anything, cause the Displays are resetting themselves again and again.

I checked the STC power supply as well, so +5, +3.3 and +12 Vdc are present on the Axial Drive board.

I ran out of ideas on how to solve this issue, I checked the connectors going to the table and going to the STC backplane and even disconnected them and plugged them back up, with no success. If you have any idea, let me know, please.

Thanks.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: LightSpeed 16 Error codes

March 31, 2016 12:17

a 8A (F5) fuse blew up in the HEMRC module, so it will get replaced, but I still have the issue of the Gantry display reseting itself everytime, and it doesn't let me control the table to make any exposures.

1 person liked this
Log inor Register

to rate and post a comment

(6)

Nawaz Salman

re: LightSpeed 16 Error codes

November 29, 2021 10:11

GE 16 slice xtrem console....image not recon RAC HARDWARE FAILURE ....wat s that problem .....can u help....

Log inor Register

to rate and post a comment

(6)
(5)

Camilo Rodriguez

re: re: LightSpeed 16 Error codes

January 07, 2022 11:11

This may be a problem with the RAC board of the IG computer. Run racdiags from console and check the results.